This forum is intended for questions about kinetics, Surface Plasmon Resonance and the instruments related to these techniques.
problems with Response unit
- Zhumabekova
- Topic Author
- Visitor
3 years 2 weeks ago #1
by Zhumabekova
problems with Response unit was created by Zhumabekova
Dear All, hello. Need urgent help! How to check what's going on inside the Biacore machine? The response unit for the old and new chip showed 2000 RU, whereas before it showed 25000 RU. What could be the reasons for this? Thank you.
Please Log in or Create an account to join the conversation.
- Arnoud
- Visitor
3 years 2 weeks ago - 3 years 2 weeks ago #2
by Arnoud
Replied by Arnoud on topic problems with Response unit
Hi, can you tell the instrument type you use?
Check if the sensor chip is properly inserted in the sheeting. Which kind of sensor chip did you use and was it new?
When you take is out, are both surface sides clean?
To check proper working of the instrument, dock a system check sensor chip, switch to fresh HBS buffer and normalize the detector before running the system check according to the procedure.
kind regards
Arnoud
Check if the sensor chip is properly inserted in the sheeting. Which kind of sensor chip did you use and was it new?
When you take is out, are both surface sides clean?
To check proper working of the instrument, dock a system check sensor chip, switch to fresh HBS buffer and normalize the detector before running the system check according to the procedure.
kind regards
Arnoud
Last edit: 3 years 2 weeks ago by Arnoud.
Please Log in or Create an account to join the conversation.
- Zhumabekova
- Topic Author
- Visitor
3 years 2 weeks ago #3
by Zhumabekova
Replied by Zhumabekova on topic problems with Response unit
Thank you.
Instrument: Biacore X100
Sensor chip: CM5 (tried old and new)
Yes, both surface sides clean
During the system check, response and injections failed, peristaltic pump calibration - not tested, and only noise passed.
Instrument: Biacore X100
Sensor chip: CM5 (tried old and new)
Yes, both surface sides clean
During the system check, response and injections failed, peristaltic pump calibration - not tested, and only noise passed.
Please Log in or Create an account to join the conversation.
- Arnoud
- Visitor
3 years 2 weeks ago #4
by Arnoud
Replied by Arnoud on topic problems with Response unit
Hi,
You can check if the flow is correct: no leakage anywhere and check if the temperature is properly maintained. Other than that I don't have a solution for you.
I recommend to contact your sales representative to solve this.
Arnoud
You can check if the flow is correct: no leakage anywhere and check if the temperature is properly maintained. Other than that I don't have a solution for you.
I recommend to contact your sales representative to solve this.
Arnoud
Please Log in or Create an account to join the conversation.
Moderators: Arnoud, Arnoud