Hi,
Analyte quantification is always surrounded with some variation. When you measure the concentration based on the UV (A280) you should use the molar extinction coefficient of that analyte to calculate the concentration. For an antibody this is around 1.4 (E 1 mg/ml / A280 nm) but variations occur.
The amino acid analysis depends on a calibration curve and possibly has its own region of variation.
I can imagine that when you do a Bradford or BCA protein concentration assay that you also will find slight variations in concentration depending on the calibration you use.
Bottom line is that a thorough exact concentration determination is very difficult.
There is a PDF document about this on the SPR-pages
www.sprpages.nl/downloads/how-to#Tprotein
As for your results I would say that that the k
a is around 1E6 M
-1s
-1 (2 significant digits is too much) and the K
D around a 100 pM.
Kind regards Arnoud