Hi,
Equilibration means that the system is washed thoroughly with the running buffer. If you run a sensorgram the drift in the lines should be almost zero indicating a stable system. Run standby, prime several times or perform a manual run for some time and check the lines.
When you run a method it is advised to program 3-5 dummy cycles to 'equilibrate' and stabilize the system. The dummy runs consist of all steps but without the analyte and including the regeneration step and extra washes is needed.
In a cycle you can program a wait period of 5 minutes to stabilize the curves after starting the sensorgram. Generally the system starts from a low flow rate and the experiment is at a higher flow rate. The wait period will stabilize the curves.
- prime after changing a sensor chip or flow buffer bottle
- run a sensorgram until the curves are stable
- program a wait period in a cycle and dummy cycles each time you run a program
Arnoud