SPR-Pages

pagelogo
back
Calibration Free concentration analysis
back
This forum is intended for questions about kinetics, Surface Plasmon Resonance and the instruments related to these techniques.

Calibration Free concentration analysis

  • mtarca
  • Topic Author
  • Visitor
  • Visitor
6 years 4 months ago #1 by mtarca
Dear SPR Pages,
Can someone explain what is Calibration Free concentration analysis? How does it get concentration as opposed to using a UV spec or a Bradford technique in determining protein concentration? Also, how would it compare? If I got 1 mg/mL using a UV spec. Would I this CFCA to be close to 1 mg/mL also? Kindly comment.

Thanks.

Please Log in or Create an account to join the conversation.

  • Arnoud
  • Visitor
  • Visitor
6 years 4 months ago #2 by Arnoud
Replied by Arnoud on topic Calibration Free concentration analysis
Calibration Free Concentration Analysis (CFCA) (1, 4, 5) (2, 6) makes it possible the measure the active concentration of a ligand without a calibration curve. The method makes use of the mass transport limitation (km) calculated from the molecular mass (Mw), which occurs when high-density ligand surfaces are used. By injecting the analyte at two different flow rates (e.g. 10 – 90 µl min-1), the active analyte concentration can be calculated from the slopes of the curves. The advantage of the CFCA over more traditional concentration measurements is that CFCA measures the ‘active’ concentration. This means that only functional compounds are measured even in a crude solution. However, not all interactions are suitable for CFCA. The interaction should have a rapid association (ka > 5 x 105 M-1s-1) and have a reasonable affinity (KD < 10-7 M). The method assumes a 1:1 interaction and the molecular weight of the interactant must be known to calculate the diffusion coefficient. The accuracy is mainly dependent on the diffusion coefficient of the analyte but compared to other concentration measurement the CFCA method is very accurate (3).

Kinds regards
Arnoud


1. Chavane, N., Jacquemart, R., Hoemann, C. D., et al.; At-line quantification of bioactive antibody in bioreactor by surface plasmon resonance using epitope detection. Analytical Biochemistry (378) 2: 158-165; 2008.
2. Christensen, L. L.; Theoretical analysis of protein concentration determination using biosensor technology under conditions of partial mass transport limitation. Analytical Biochemistry (249) 2: 153-164; 1997.
3. Helmerhorst, E., Chandler, D. J., Nussio, M., et al.; Real-time and Label-free Bio-sensing of Molecular Interactions by Surface Plasmon Resonance: A Laboratory Medicine Perspective. The Clinical Biochemist Reviews (33) 4: 161-173; 2012.
4. Pol, E., Karlsson, R., Roos, H., et al.; Biosensor-based characterization of serum antibodies during development of an anti-IgE immunotherapeutic against allergy and asthma. Journal of Molecular Recognition (20) 1: 22-31; 2007.
5. Pol, E., Roos, H., Markey, F., et al.; Evaluation of Calibration-Free Concentration Analysis provided by Biacore™ systems. Analytical Biochemistry 2016.
6. Sigmundsson, K., Masson, G., Rice, R., et al.; Determination of Active Concentrations and Association and Dissociation Rate Constants of Interacting Biomolecules: An Analytical Solution to the Theory for Kinetic and Mass Transport Limitations in Biosensor Technology and Its Experimental Verification. Biochemistry (41) 26: 8263-8276; 2002.

Please Log in or Create an account to join the conversation.

Moderators: ArnoudArnoud
Powered by Kunena Forum