SPR-Pages

pagelogo
back
Please help me to analyze the experimen
back
This forum is intended for questions about kinetics, Surface Plasmon Resonance and the instruments related to these techniques.

Please help me to analyze the experimental result !!!

  • Wang
  • Topic Author
  • Visitor
  • Visitor
6 years 7 months ago #1 by Wang
dear arnoud,
I recently made a kinetic experiment on antibodies, and I'm not sure if my results are reliable. Please help me analyze them.
I use CM5 chip, immobilized anti human IgG Fc, ligand is the antibody of human source, the analyte is protein.
The antibody's dissociatio is very slow, and the HPLC-SEC graph shows that the purity of ligand is 91%, 9% of the polymers, and this will slow down the dissociation of the analyte?
What kind of optimization is needed for this experiment?
Thank you a lot.
Attachments:

Please Log in or Create an account to join the conversation.

  • Arnoud
  • Visitor
  • Visitor
6 years 7 months ago #2 by Arnoud
Hi,
The picture is not that good but you have a long dissociation and that is good for a strong interaction. The association time may be a bit longer so the association has some curvature for a better ka estimation.
What you can do is centrifuge the analyte at 15000g for 5 - 10 minutes to remove aggregates / polymers.
I don't think that de polymers are the source of a slow dissociation because 91% is monomer. It seems that you have a strong interaction.

Kind regards
Arnoud

Please Log in or Create an account to join the conversation.

Moderators: ArnoudArnoud
Powered by Kunena Forum