SPR-Pages

pagelogo
back
Sensorgram with a grey background - SPRp
back
This forum is intended for questions about kinetics, Surface Plasmon Resonance and the instruments related to these techniques.

Sensorgram with a grey background

  • zhaojing0201
  • Topic Author
  • Visitor
  • Visitor
6 years 9 months ago #1 by zhaojing0201
Sensorgram with a grey background was created by zhaojing0201
Hi everyone,

I'm working on protein interaction using a BIAcore 3000 system.
Today when I use the machine. I saw my Sensorgram with a grey background when I choose single flow cell, the value was pretty high. After I dock my CM5 chip, the signal of 2-1 3-1 and 4-1 was -75000 around.

Dose anyone know what's the problem?

Thanks!
Jakcie.

Please Log in or Create an account to join the conversation.

  • Arnoud
  • Visitor
  • Visitor
6 years 9 months ago - 6 years 9 months ago #2 by Arnoud
Replied by Arnoud on topic Sensorgram with a grey background
Hi Jackie,

In general this is a sign that the optical system is not working properly.
In most cases there is air on the sensor chip. Or you docked a maintenance chip!

First prime to flush out the air and run a sensorgram again.
Not good. Undock and inspect the sensor chip. It should be clean on both sides.
Dock again and run a sensorgram. It should be ok now.
If not you can try another chip or you have to call in the maintenance engineer.

Kind regards.
Last edit: 6 years 9 months ago by Arnoud. Reason: typo

Please Log in or Create an account to join the conversation.

  • zhaojing0201
  • Topic Author
  • Visitor
  • Visitor
6 years 9 months ago #3 by zhaojing0201
Replied by zhaojing0201 on topic Sensorgram with a grey background
Thanks for your quick reply!!!

I ran Prime, Flush, Unclogging and desorb using a maintainace chip and then check my own CM5 chip. The Signal of most flow cell become normal but it's still about twice higher than what I get last time. Dose this mean there is still some air in the system?

Thanks!
Jackie.

Please Log in or Create an account to join the conversation.

  • zhaojing0201
  • Topic Author
  • Visitor
  • Visitor
6 years 9 months ago #4 by zhaojing0201
Replied by zhaojing0201 on topic Sensorgram with a grey background
Hi Arnoud,

When you said inspect the chip. Did you mean check the golden surface?
I took a picture of the surface of my chip. It seems there are something stuck on it.

Would this be a problem?

Thanks so much!

Jackie.
Attachments:

Please Log in or Create an account to join the conversation.

  • Arnoud
  • Visitor
  • Visitor
6 years 9 months ago - 6 years 9 months ago #5 by Arnoud
Replied by Arnoud on topic Sensorgram with a grey background
The gold surface should be perfectly clean on both sides.
This looks as if it is some buffer residue.

You can try to wash the surface and dry both sides with a clean tissue. Dry only the sides of the gold surface and dry also the glass side. You can also use air or a stream of nitrogen to dry the surface.

Depending on what is on the chip my experience is that a baseline with HBS is within 14000 - 25000 RU.
Last edit: 6 years 9 months ago by Arnoud.

Please Log in or Create an account to join the conversation.

  • zhaojing0201
  • Topic Author
  • Visitor
  • Visitor
6 years 9 months ago #6 by zhaojing0201
Replied by zhaojing0201 on topic Sensorgram with a grey background
Thanks Arnoud,

I never know that I can wash the surface by hand. :woohoo:
I'm a little worried if I would wash away any immobilized ligand during washing.
Is this a problem?

Thanks!

Jackie.

Please Log in or Create an account to join the conversation.

Moderators: ArnoudArnoud
Powered by Kunena Forum