Bubbles in the system
- papillon
- Topic Author
- New Member
-
Less
More
- Thank you received: 0
7 years 10 months ago #1
by papillon
Bubbles in the system was created by papillon
Hi all,
We have make a mistake and we have selected the wrong injection place. There were not analyte but an empty place. Then air enter in the system. How is the best wat to purge the system in order to eliminate the bubbles that can appear in the next sensograms.
Do you think that the bubbles in the sensograms could be a problem in order of obtain the kinetic constants or concentrations?
Thank you everybody.
Papillon
We have make a mistake and we have selected the wrong injection place. There were not analyte but an empty place. Then air enter in the system. How is the best wat to purge the system in order to eliminate the bubbles that can appear in the next sensograms.
Do you think that the bubbles in the sensograms could be a problem in order of obtain the kinetic constants or concentrations?
Thank you everybody.
Papillon
Please Log in or Create an account to join the conversation.
- Arnoud
- Moderator
-
Less
More
- Thank you received: 0
7 years 10 months ago #2
by Arnoud
Replied by Arnoud on topic Bubbles in the system
Hi Papillon,
In general, a prime or a flow at the highest flow rate should push air bubbles out the system.
You can monitor the baseline to observe that it stable without any spikes of air.
If you have small air bubbles in the sensorgram, say 1-3 seconds, you can mask them or remove them from the sensorgram. If there are only a few missing points from the sensorgram, the evaluation software will work properly.
regards
Arnoud
In general, a prime or a flow at the highest flow rate should push air bubbles out the system.
You can monitor the baseline to observe that it stable without any spikes of air.
If you have small air bubbles in the sensorgram, say 1-3 seconds, you can mask them or remove them from the sensorgram. If there are only a few missing points from the sensorgram, the evaluation software will work properly.
regards
Arnoud