How can I solute reference channel's curve?
- Koreanraichu
- Topic Author
- Visitor
7 years 9 months ago #1
by Koreanraichu
(run at 17th Jan)
(run at same day)
Dotted graph is reference channel, and other line is sample. sample channel changes(buffer-sample tube-buffer) when checked equilibrium, but reference don't changes tube or other solution, it keep going with running buffer until experiment end. But as you see above graph, reference channel have a curve... what should I do in this case?
How can I solute reference channel's curve? was created by Koreanraichu
(run at 17th Jan)
(run at same day)
Dotted graph is reference channel, and other line is sample. sample channel changes(buffer-sample tube-buffer) when checked equilibrium, but reference don't changes tube or other solution, it keep going with running buffer until experiment end. But as you see above graph, reference channel have a curve... what should I do in this case?
Please Log in or Create an account to join the conversation.
- Arnoud
- Visitor
7 years 9 months ago #2
by Arnoud
Replied by Arnoud on topic How can I solute reference channel's curve?
I tried to look up the type of instrument you have but could not find it. Do have a reference.
If I understand correctly, the reference is just running buffer only. If you get a peak in the reference, something is wrong with the instrument (e.g. not clean, a bad valve not switching correctly, pump not stable).
You cannot use these curves.
First try to do some dummy experiments with buffer injections until all injections give perfectly flat curve. Then inject buffer with 25 mM extra NaCl to mimic bulk effect. The curves should jump up and get stable during the injection and jump down to baseline after injection. If not check your system.
Kind regards
Arnoud
If I understand correctly, the reference is just running buffer only. If you get a peak in the reference, something is wrong with the instrument (e.g. not clean, a bad valve not switching correctly, pump not stable).
You cannot use these curves.
First try to do some dummy experiments with buffer injections until all injections give perfectly flat curve. Then inject buffer with 25 mM extra NaCl to mimic bulk effect. The curves should jump up and get stable during the injection and jump down to baseline after injection. If not check your system.
Kind regards
Arnoud
Please Log in or Create an account to join the conversation.
- Koreanraichu
- Topic Author
- Visitor
7 years 9 months ago #3
by Koreanraichu
Replied by Koreanraichu on topic How can I solute reference channel's curve?
All of progress(streptavidin-linker-sample), in SA and linker, reference curve saw clear, and no get out of range and shake. but it was only in DNA injection,
You're understand correctly, dotted line is reference channel with running buffer only.
You're understand correctly, dotted line is reference channel with running buffer only.
Please Log in or Create an account to join the conversation.
- Koreanraichu
- Topic Author
- Visitor
7 years 9 months ago #4
by Koreanraichu
Replied by Koreanraichu on topic How can I solute reference channel's curve?
I did experiment that you advised, reference curve is flat, has no problem.. I think that is buffer's problem.
Please Log in or Create an account to join the conversation.