This forum is intended for questions about kinetics, Surface Plasmon Resonance and the instruments related to these techniques.

Long spike after injection

  • wagquint
  • Topic Author
  • Visitor
  • Visitor
8 years 4 months ago #1 by wagquint
Long spike after injection was created by wagquint
I am trying to perform a single cycle kinetics to evaluate an antibody against a toxin. The toxin is immobilized in a CM5 sensorchip by amine coupling. Everything could be right except by the fact that just after the sample injection there is a long spike that happens during a few seconds, which at the end prevents the calculation by Biacore software. I changed the buffer, I did desorb/sanitize and superclean, I prepared another sensorchip... Could anyone tell me what else I should do?

An example of a sensorgram:

Attachment sensorgram1.png not found



Zoom in:

Attachment sensorgram2.png not found

Attachments:

Please Log in or Create an account to join the conversation.

  • Arnoud
  • Visitor
  • Visitor
8 years 4 months ago #2 by Arnoud
Replied by Arnoud on topic Long spike after injection
Hi,

What kind of Biacore do you use? This happens already several times? And why not at the last three injections? I think it is at the start of the injections. Did you check if there is no air in the tubes?

Kind regards

Arnoud

Please Log in or Create an account to join the conversation.

  • wagquint
  • Topic Author
  • Visitor
  • Visitor
8 years 4 months ago #3 by wagquint
Replied by wagquint on topic Long spike after injection
Hi, I am using Biacore T200.
Yes, it happens most but not all the runs. I thought it could be air bubbles, but this means I may have equipment problems...

Thank you

Please Log in or Create an account to join the conversation.

  • Arnoud
  • Visitor
  • Visitor
8 years 4 months ago #4 by Arnoud
Replied by Arnoud on topic Long spike after injection
What you can check is the pumps, behind the door.
If you trigger a prime you can see if the pumps are fully filled while operating.
If they draw some air you know you need some servicing.

Arnoud

Please Log in or Create an account to join the conversation.

  • wagquint
  • Topic Author
  • Visitor
  • Visitor
8 years 4 months ago - 8 years 4 months ago #5 by wagquint
Replied by wagquint on topic Long spike after injection
Hi!
I called GE service. They tightened the degasser connectors and advised me to perform "Flush system wash". I did it twice and the problem seems solved.
Thank you again! B)
Last edit: 8 years 4 months ago by wagquint.

Please Log in or Create an account to join the conversation.

Moderators: ArnoudArnoud