small oscillation in the beginning of the sensorgram

  • jonbuitb
  • Topic Author
  • Visitor
  • Visitor
8 years 9 months ago - 8 years 9 months ago #1 by jonbuitb
Hello,

I would like to now why I find this small oscillations in the beginning of the curves for the high concentration of analyte (500 uM and 1 mM) in the sensorgram. the protein is on the chip and the small molecule is injected (in the same running buffer). The stock of the small molecule was in 100% DMSO.

Experiment condition:

The experiment was done in a Biacore T200 at 25C.
Running buffer : 10 mM HEPES pH 7.4, 150 mM NaCL, 0.05% Tween.
Small molecule dilution:Running buffer plus 5%DMSO (1mM), 2.5%DMSO (500 uM), 1,25 (250 uM) and go on. Could be these peaks have a DMSO origin?


Thanks.
Attachments:
Last edit: 8 years 9 months ago by jonbuitb.

Please Log in or Create an account to join the conversation.

  • Arnoud
  • Visitor
  • Visitor
8 years 9 months ago - 8 years 9 months ago #2 by Arnoud
Hi Jon,

Looking at the picture I suspect that there is a significant DMSO buffer jump in the original curves, am I right?
In addition the channels are in series which means that the buffer jump is not exactly at the same time. And when subtracting this negative of positive jumps can occur.
You can try to align the curves better or to try to minimize the buffer jumps (best option).


Kind regards

Arnoud