Scrubber2 reference question
- jenna
- Topic Author
- Visitor
8 years 10 months ago #1
by jenna
Scrubber2 reference question was created by jenna
Hi! When working in Scrubber2, there is an option under the "Reference" tab to indicate which flow cell is the reference, and then also to subtract either curve or jump. What does it mean to subtract the curve or the jump? Is there any difference in the result? Thanks!
Please Log in or Create an account to join the conversation.
- Arnoud
- Visitor
8 years 10 months ago #2
by Arnoud
Replied by Arnoud on topic Scrubber2 reference question
Hi Jenna,
When selecting for curve, the whole reference curve is used. The advantage is, that when the curve has a waobbled uring injection, this is also subtracted.
When selecting for jump, you have to position the blue line at the end of the injection jump. The response value at the line will be subtracted from the other curves. The shape of the curves is not changed only the response.
Kind regards
Arnoud
When selecting for curve, the whole reference curve is used. The advantage is, that when the curve has a waobbled uring injection, this is also subtracted.
When selecting for jump, you have to position the blue line at the end of the injection jump. The response value at the line will be subtracted from the other curves. The shape of the curves is not changed only the response.
Kind regards
Arnoud
Please Log in or Create an account to join the conversation.
- jenna
- Topic Author
- Visitor
8 years 10 months ago #3
by jenna
Replied by jenna on topic Scrubber2 reference question
Hi Arnoud, thanks for your quick response! What do you mean by "injection jump"? I am working with determining the off-rate of serum antibodies (specific to immobilized antigen) -- is one method "better" than the other? Thanks again! Best, Jenna
Please Log in or Create an account to join the conversation.
- Arnoud
- Visitor
8 years 10 months ago - 8 years 10 months ago #4
by Arnoud
Replied by Arnoud on topic Scrubber2 reference question
Hi Jenna,
With injection jump i mean the bulk jump you can have when injecting the analyte. When you have matched the flow and analyte buffer you hardly see it. You apply the 'jump' reference subtraction method with the values at the end of the analyte injection.
In your case it makes no difference since you do off-rate ranking.
kinds regards
Arnoud
With injection jump i mean the bulk jump you can have when injecting the analyte. When you have matched the flow and analyte buffer you hardly see it. You apply the 'jump' reference subtraction method with the values at the end of the analyte injection.
In your case it makes no difference since you do off-rate ranking.
kinds regards
Arnoud
Last edit: 8 years 10 months ago by Arnoud.
Please Log in or Create an account to join the conversation.