Sensorgram increase at the start
- skuha
- Topic Author
- Offline
- Banned
Less
More
- Posts: 17
- Thank you received: 0
8 years 10 months ago #1
by skuha
Sensorgram increase at the start was created by skuha
Please Log in or Create an account to join the conversation.
- Arnoud
- Visitor
8 years 10 months ago #2
by Arnoud
Replied by Arnoud on topic Sensorgram increase at the start
Hi Ahmed,
The spike come from the fact that the channels are in series. When subtracting a reference, both channels are a little out of phase.
In the Biacore you can use the direct referencing during the experiment which will give less spikes.
Or you can adjust the injection time point t=0 for each curve.
When using a program like Scrubber, you kan artificially remove the spikes.
Large spikes occur mainly when the bulk effect is large (>25 RU) and the compensation by the reference channel is not optimal.
Kind regards
Arnoud
The spike come from the fact that the channels are in series. When subtracting a reference, both channels are a little out of phase.
In the Biacore you can use the direct referencing during the experiment which will give less spikes.
Or you can adjust the injection time point t=0 for each curve.
When using a program like Scrubber, you kan artificially remove the spikes.
Large spikes occur mainly when the bulk effect is large (>25 RU) and the compensation by the reference channel is not optimal.
Kind regards
Arnoud