See, i do understand that RI is for bulk shift accommodation when the running buffer and sample buffers are mismatched. but my question is when I chose a 1:1 model of curved fitting what criteria do i also look into to set RI=0 ? or i use the default setting. ? My sample buffer and running buffer is the same...
Please Log in or Create an account to join the conversation.
- Arnoud
- Visitor
9 years 7 months ago #2
by Arnoud
Replied by Arnoud on topic when should we set RI=0 for a 1:1 fit?
In general the RI is fitted along with the other parameters. Fitted RI should be low and proprtional to the analyte concentration.
I my opinion the RI is often calculated too large.
I always start with RI=0 to determine the Rmax, ka and kd. Then I add RI-fitting and start the fitting with the values of the previous fitting.
Look at the second figure at www.sprpages.nl/sensorgram-tutorial/curve-fitting.html .
Make sure you don't have this.
I my opinion the RI is often calculated too large.
I always start with RI=0 to determine the Rmax, ka and kd. Then I add RI-fitting and start the fitting with the values of the previous fitting.
Look at the second figure at www.sprpages.nl/sensorgram-tutorial/curve-fitting.html .
Make sure you don't have this.
Please Log in or Create an account to join the conversation.
- BrukerSPR
- Topic Author
- Offline
- Junior Member
Less
More
- Posts: 22
- Thank you received: 0
9 years 7 months ago #3
by BrukerSPR
Replied by BrukerSPR on topic when should we set RI=0 for a 1:1 fit?
hey arnoud,
I do have large RI jumps after setting RI=0 and not as large when I have RI as a fitted parameter...
Any inputs?
I do have large RI jumps after setting RI=0 and not as large when I have RI as a fitted parameter...
Any inputs?
Please Log in or Create an account to join the conversation.
- Arnoud
- Visitor
9 years 7 months ago - 9 years 7 months ago #4
by Arnoud
Replied by Arnoud on topic when should we set RI=0 for a 1:1 fit?
Lets clear something up.
If I use RI=0 I mean that I keep RI as a constant on zero. Thus assuming there is no RI effect. Ergo there can be no RI value other than zero.
If you set RI=0 and the fit it locally (must be local!) then the progam wil fit a value starting with zero. The default fittings use often a starting value of 10 or 50. When fitting clean sensorgram the answers should be comparable regardless the starting value (unless really extreme).
So I normally start fitting without RI --> RI=0 and a constant, Then after I have the ka, kd and Rmax, I fill in these values and set RI=0 and fit locally.
Maybe you can show the two fittings with values (or otherwise send a message via the contact form).
If I use RI=0 I mean that I keep RI as a constant on zero. Thus assuming there is no RI effect. Ergo there can be no RI value other than zero.
If you set RI=0 and the fit it locally (must be local!) then the progam wil fit a value starting with zero. The default fittings use often a starting value of 10 or 50. When fitting clean sensorgram the answers should be comparable regardless the starting value (unless really extreme).
So I normally start fitting without RI --> RI=0 and a constant, Then after I have the ka, kd and Rmax, I fill in these values and set RI=0 and fit locally.
Maybe you can show the two fittings with values (or otherwise send a message via the contact form).
Last edit: 9 years 7 months ago by Arnoud.
Please Log in or Create an account to join the conversation.
- BrukerSPR
- Topic Author
- Offline
- Junior Member
Less
More
- Posts: 22
- Thank you received: 0
9 years 7 months ago #5
by BrukerSPR
Replied by BrukerSPR on topic when should we set RI=0 for a 1:1 fit?
there is some undercurrent of confusion that I have... can i talk to you?. any which way i will post the sensogram ... I use Biaevaluation software t100/200..
Please Log in or Create an account to join the conversation.
- BrukerSPR
- Topic Author
- Offline
- Junior Member
Less
More
- Posts: 22
- Thank you received: 0
9 years 7 months ago #6
by BrukerSPR
Replied by BrukerSPR on topic when should we set RI=0 for a 1:1 fit?
here i paste two sensograms... one with RI=0 and one with RI= fitted..
I have fcgRIIIa immobilised at 250 RU
there are 8 conc. of the drug product
Query is, I see when I use RI=0 the fitting is bad in 1:1 model and when I use RI= fitted i see a better fit... but i don't know if i could use RI=fitted or when I should use RI= 0...
I have fcgRIIIa immobilised at 250 RU
there are 8 conc. of the drug product
Query is, I see when I use RI=0 the fitting is bad in 1:1 model and when I use RI= fitted i see a better fit... but i don't know if i could use RI=fitted or when I should use RI= 0...
Please Log in or Create an account to join the conversation.