SPR-Pages

pagelogo
back
Low protein signal during immobilisation
back
This forum is intended for questions about kinetics, Surface Plasmon Resonance and the instruments related to these techniques.

Low protein signal during immobilisation

  • biber
  • Topic Author
  • Visitor
  • Visitor
9 years 8 months ago #1 by biber
Hİ everybody,

I use Bionavi210A SPR. I'm still learning. I try to immobilze HSA with amino coupling on CMD (=CM5) chip. But my protein signal is very low. My coupling buffer is 10 mM Na Acetate pH: 5.2. HSA is globulin and fatty acid free. I tried a lot of different chemical such as different protein (BSA), different EDC and NHS, Different buffer ( 5 mM MES pH: 5.0) but nothing changes. I also tried different concentraion of proteins. I sent you my sensogram results.





I sent you my sensogram results for BSA and HSA .

Do you have any recommadation to solve this problem?
Attachments:

Please Log in or Create an account to join the conversation.

  • MQAdmin03
  • Visitor
  • Visitor
9 years 8 months ago #2 by MQAdmin03
Replied by MQAdmin03 on topic Low protein signal during immobilisation
Hi Biber,

Not sure what i see in your sensorgrams (immobilization or experiment; please explain) but i have some suggestions.
- take a fresh equilibrated sensor chip
- inject the HSA solution in different pH buffers ( www.sprpages.nl/immobilization/ligand-pre-concentration.html )
- take the pH solution with the "best" pre concentration properties.
- then make fresh NHS and EDC according to the Biacore protocol or by the provided BioNavis protocol.
- Try again.

Please Log in or Create an account to join the conversation.

Moderators: Arnoud
Powered by Kunena Forum