SPR dissociation curve deviation
- OldForum
- Topic Author
- Visitor
13 years 8 months ago #1
by OldForum
SPR dissociation curve deviation was created by OldForum
I have just started with SPR to detect interaction between two proteins. I am using CM5 sensor chips for the immobilization of one of my protein (ligand protein)and flowing another analyte protein (Molecular weight 20KD) through the flow channel. I am getting good immobilization of my ligand protein as detected by resonance units and a good enough resonance units obtained during analyte injection to account for interaction.Though the association is good enough to account for binding but the dissociation is not showing as expected.the dissociation curve is not falling after the interaction has happened.Although, I have already checked through in vitro experiments that these two proteins interact with each other. What could be the possible reason for the dissociation curve not falling as it should be?and What should I do next to optimize the interaction between my two proteins using SPR to show an usual and typical dissociation curve?
Ashok
Ashok
Please Log in or Create an account to join the conversation.
- OldForum
- Topic Author
- Visitor
13 years 8 months ago #2
by OldForum
Replied by OldForum on topic SPR dissociation curve deviation
Hi,
Has the binding response a nice shape? Could it be non-specific binding what you are seeing?
JoseAntonio
Has the binding response a nice shape? Could it be non-specific binding what you are seeing?
JoseAntonio
Please Log in or Create an account to join the conversation.
- OldForum
- Topic Author
- Visitor
13 years 8 months ago #3
by OldForum
Replied by OldForum on topic SPR dissociation curve deviation
So you have no dissociation after the analyte has passed the ligand?
The dissociation, is it a flat line and there is no dissociation at all? Because no dissociation at all is fishy.
Can you regenerate the ligand and repeat the interaction? If so, the interaction is possibly strong and a slow dissociation can be expected.
You probably checked already the workings of the machine confirming that the buffer is flowing over the ligand after the sample plug.
Regards
Arnoud
The dissociation, is it a flat line and there is no dissociation at all? Because no dissociation at all is fishy.
Can you regenerate the ligand and repeat the interaction? If so, the interaction is possibly strong and a slow dissociation can be expected.
You probably checked already the workings of the machine confirming that the buffer is flowing over the ligand after the sample plug.
Regards
Arnoud