RU decreasing during injection
- OldForum
- Topic Author
- Visitor
13 years 8 months ago #1
by OldForum
RU decreasing during injection was created by OldForum
Hi,
Did anyone have such experience? When you inject a sample, it will increase immediately and go to the plateau then after half of the injection, it starts to decrease (but the injection is not over). That is really crazy. The RU is not low, they are not noise. I tried several different concentration, they are always like this. I also tried to use short injection time(less injection volumne), but they are the same.
Did any one of you have similar problem? How did you solve the problem.
If someone could help me, that will be very helpful for me.
Did anyone have such experience? When you inject a sample, it will increase immediately and go to the plateau then after half of the injection, it starts to decrease (but the injection is not over). That is really crazy. The RU is not low, they are not noise. I tried several different concentration, they are always like this. I also tried to use short injection time(less injection volumne), but they are the same.
Did any one of you have similar problem? How did you solve the problem.
If someone could help me, that will be very helpful for me.
Please Log in or Create an account to join the conversation.
- OldForum
- Topic Author
- Visitor
13 years 8 months ago #2
by OldForum
Replied by OldForum on topic RU decreasing during injection
Hi,
We see it sometimes. The exact cause can be different.
1) injection system: by mixing between sample and flow buffer during injection the concentration of analyte is not constant (lower). use KInject and check the injection system. Clean injection loop and make sure the pumps are working correctly.
2) sample: it is possible that the sample is not homogenous. The larger molecules seems more abundant and bind first. The smaller molecules however have a higher affinity and will displace the larger molecules lowering the signal. Make a PAGE gel to look for the purity of the analyte.
3) ?: we sometimes can not find a cause....
regards
Arnoud
We see it sometimes. The exact cause can be different.
1) injection system: by mixing between sample and flow buffer during injection the concentration of analyte is not constant (lower). use KInject and check the injection system. Clean injection loop and make sure the pumps are working correctly.
2) sample: it is possible that the sample is not homogenous. The larger molecules seems more abundant and bind first. The smaller molecules however have a higher affinity and will displace the larger molecules lowering the signal. Make a PAGE gel to look for the purity of the analyte.
3) ?: we sometimes can not find a cause....
regards
Arnoud
Please Log in or Create an account to join the conversation.
- OldForum
- Topic Author
- Visitor
13 years 8 months ago #3
by OldForum
Replied by OldForum on topic RU decreasing during injection
Something like that has happened to me. In my case the cause was that I hadnt put Tween in the sample.