Regeneration of Biosensor chip

  • OldForum
  • Topic Author
  • Visitor
  • Visitor
17 years 4 months ago #1 by OldForum
Regeneration of Biosensor chip was created by OldForum
Hello,

I am using a CM5 chip and use ligand thiol coupling for immobilizing the ligand. Can someone tell me the way to remove the originally bound ligand and reuse the chip. As per the protocol mentioned in the SPR pages it asks us to flow 0.1M DTT in borate buffer. Can someone please tell me the Concenmtration and the pH of the borate buffer to be employed? Thanks for the help

Regards,
Satyen

Please Log in or Create an account to join the conversation.

  • OldForum
  • Topic Author
  • Visitor
  • Visitor
17 years 4 months ago #2 by OldForum
Replied by OldForum on topic Regeneration of Biosensor chip
Hi, I would take the same borate buffer as in the immobilization, i.e. 0.1 M borate pH 8.0.

The reference on the SPR-pages seems nothing to do with surface regeneration...

Maybe this artikle will be of help:

Reversible Surface Thiol Immobilization of Carboxyl Group
Containing Haptens to a BIAcore Biosensor Chip Enabling Repeated
Usage of a Single Sensor Surface
Ulrich Schlecht, Yoko Nomura, Till Bachmann, and Isao Karube*
Bioconjugate Chem. 2002, 13, 188193


Arnoud