These are the posts of the old forum. It was not possible to transfer the user data, so they are missing in most of the posts. For new questions, go to the general discussions.
Unspecific signals
- OldForum
- Topic Author
- Visitor
18 years 8 months ago #1
by OldForum
Unspecific signals was created by OldForum
Hi all,
Lastly I am seeing a lot of unspecific signals like interactions that I did not hope "a priori". Some of these occurs with streptavidin immobilized over the chip.
Is this a normal behaviour?
Usually NaCl buffer concentration is 150 mM and 0.005%Tween20.
Also, only comercial buffer injections show that the signal increases as the injection proceeds.
Any suggestion will be very appreciated
Thanks a lot
Lastly I am seeing a lot of unspecific signals like interactions that I did not hope "a priori". Some of these occurs with streptavidin immobilized over the chip.
Is this a normal behaviour?
Usually NaCl buffer concentration is 150 mM and 0.005%Tween20.
Also, only comercial buffer injections show that the signal increases as the injection proceeds.
Any suggestion will be very appreciated
Thanks a lot
Please Log in or Create an account to join the conversation.
- OldForum
- Topic Author
- Visitor
18 years 8 months ago #2
by OldForum
Replied by OldForum on topic Unspecific signals
How does your non-specific signal looks like? Is a gradual drift up or down or jumps?
Is it on all channels the same?
Which commercial buffers do ou use? Do you repack them e.g. in an other flask?
Arnoud
Is it on all channels the same?
Which commercial buffers do ou use? Do you repack them e.g. in an other flask?
Arnoud
Please Log in or Create an account to join the conversation.
- OldForum
- Topic Author
- Visitor
18 years 8 months ago #3
by OldForum
Replied by OldForum on topic Unspecific signals
Hi Arnoud, thanks for replying
The unspecific signals are like a gradual drift up in both FC1 and Fc2 channels of our Biacore X.
I do not repack commercial buffers which are from the same house as the machine is from.
I have seen a paper in which the authors increase the P20 concentration from 0.005% to 0.05% in order to minimize unspecific signals
The unspecific signals are like a gradual drift up in both FC1 and Fc2 channels of our Biacore X.
I do not repack commercial buffers which are from the same house as the machine is from.
I have seen a paper in which the authors increase the P20 concentration from 0.005% to 0.05% in order to minimize unspecific signals
Please Log in or Create an account to join the conversation.
- OldForum
- Topic Author
- Visitor
18 years 8 months ago #4
by OldForum
Replied by OldForum on topic Unspecific signals
Yes, you can raise the P20 concentration.
Sometimes a higer salt concentration can be of help but binding is often salt concentration depend. Thus you should check!
I see often drift after changing buffer type or cleaning the machine. This can last for one or two days. Thorough washing of the system is essential. Flowing overnight at 20 ul.min with some injections of the same buffer should stabilize the system.
And last. Calculate your drift back to the machine specifications. Often people want less drift than the machine can deliver.
Arnoud
Sometimes a higer salt concentration can be of help but binding is often salt concentration depend. Thus you should check!
I see often drift after changing buffer type or cleaning the machine. This can last for one or two days. Thorough washing of the system is essential. Flowing overnight at 20 ul.min with some injections of the same buffer should stabilize the system.
And last. Calculate your drift back to the machine specifications. Often people want less drift than the machine can deliver.
Arnoud
Please Log in or Create an account to join the conversation.