SPR-Pages

pagelogo
back
negative binding signals - SPRpages
back
These are the posts of the old forum. It was not possible to transfer the user data, so they are missing in most of the posts. For new questions, go to the general discussions.

negative binding signals

  • OldForum
  • Topic Author
  • Visitor
  • Visitor
19 years 6 months ago #1 by OldForum
negative binding signals was created by OldForum
Hi everybody,

Does anyone have experience in receiving negative binding signals when analyzing small molecules (500-1000Da) binding to immobilized proteins (100kDa). It would be interesting to hear about possible explanations and systems where this was seen. The sensorgrams are analyzable and show saturation binding yielding good results so the system seems to work anyhow.

As an explanation a conformational change in the protein dependent on binding could compensate the positive binding signal of the analyte binding, but who knows....

thanx Strada

Please Log in or Create an account to join the conversation.

  • OldForum
  • Topic Author
  • Visitor
  • Visitor
19 years 6 months ago #2 by OldForum
Replied by OldForum on topic negative binding signals
Can you tell me when you observe these negative binding signals? Is this already before or only after reference subtraction ?

Can you reverse the system? And so, do you see the same trange behavour ?

Please Log in or Create an account to join the conversation.

  • OldForum
  • Topic Author
  • Visitor
  • Visitor
19 years 6 months ago #3 by OldForum
Replied by OldForum on topic negative binding signals
thanx for the reply and for the very good and infomative homepage. It's very useful to have a forum to ask questions about Biacore experiments to a world wide community.

The negative sensorgrams are seen before and after referencing (double referencing). I have measure different analytes with the same result using this specific target. reversing the system has not been done yet. that would be a possibility but a difficult one because of the nature of the analyte. Our approach is to try to measure a heavier analyte to compensate for a possible conformational change or a matrix deformation yielding negative resonses.

The obtained KD values do correspond to a equilibrium assay (so there must be some kind of truth (correlation) in the biacore assay). Strangly the Rmax (or Rmin looking at the untreated sensorgrams) is dependent on how much target is immobilized.

Please Log in or Create an account to join the conversation.

  • OldForum
  • Topic Author
  • Visitor
  • Visitor
19 years 6 months ago #4 by OldForum
Replied by OldForum on topic negative binding signals
First your last remarks. Good: KD's should be verified with different set ups so your are on the right way.
Is it possible to do ITC (isothermal calorimetrie) or CD (circular diochroism). The latter is best equipped to demonstrate conformational change.

Rmax should differ with the amount of ligand immobilized because Rmax is the response by the injected analyte. Rmax depends on the Molecular weights of the ligand and analyte and the amount (binding places) of the immobilized ligand. Don't confuse Rmax with Req. Rmax is only approached by > 100x KD.

These analyte bindings are they fast (ka >10e6, kd < 10-2) or slow?

Please Log in or Create an account to join the conversation.

  • OldForum
  • Topic Author
  • Visitor
  • Visitor
19 years 6 months ago #5 by OldForum
Replied by OldForum on topic negative binding signals
the analytes shows fast association (10e5) and dissociation (10e-1). Until now, I have measured three different analytes and always seen the same negative sensorgrams and the correlation to the other assay.

ITC is planned for this system. I hope that I'll be able to confirm the results using it

Please Log in or Create an account to join the conversation.

  • OldForum
  • Topic Author
  • Visitor
  • Visitor
19 years 6 months ago #6 by OldForum
Replied by OldForum on topic negative binding signals
It is difficult to understand why your sensorgrams are negative but I think you have to look at the interaction of the solutions with your reference and active surface.

There is not much literature but this may be of help:

Karlsson R, Anal Biochem 221, 142-151 (1994)
Karlsson R, Anal Biochem 278, 1-13 (2000)

Please Log in or Create an account to join the conversation.

Powered by Kunena Forum