SPR-Pages

pagelogo
back
Linked reactions - SPRpages
back
These are the posts of the old forum. It was not possible to transfer the user data, so they are missing in most of the posts. For new questions, go to the general discussions.

Linked reactions

  • OldForum
  • Topic Author
  • Visitor
  • Visitor
19 years 3 weeks ago #1 by OldForum
Linked reactions was created by OldForum
Hi all,

I am trying to identify a conformational change upon Analyte binding to Ligand. I know that different dissociation rates with different injecting times at a constant analyte concentration is the method to resolve this question. But the A-L interaction does not reach steady-state even at high analyte concentrations (200 nM) even at long injection times.

Is there any solution?

thanks in advance and good luck with your experiments

Please Log in or Create an account to join the conversation.

  • OldForum
  • Topic Author
  • Visitor
  • Visitor
19 years 3 weeks ago #2 by OldForum
Replied by OldForum on topic Linked reactions
Try to calculate if your goal is feasable.

You can calculate the Req

Req = (ka . C / ka . C + kd) Rmax

You can estimate the time to reach equilibrium

teq = -ln(1-@)/(ka . C + kd)

where teq is the time to equilibrium and @ is the fraction of reached equilibrium, typical 95% thus 0.95

Arnoud

Please Log in or Create an account to join the conversation.

Powered by Kunena Forum